Bioanalytical MethodsBrian Booth, Ph.D.Office of Clinical PharmacologyFDA/CDER/OTS/OCPThe views expressed are those of the author and do not reflect official policy of the FDA.No official endorsement by the FDA is intended or should be inferred1

Outline1. The FDA guidance-our philosophy in therevision and our feedback2. Questions from the JBF3. FDA Revision Efforts-the Next Steps2

FDA BMV Guidance Some old issues, some new issues– Some new concepts e.g. biomarkers– Some old concepts clarified– Some things carried forward from oldGuidance– Some topics not mentioned3

FDA BMV Guidance-PhilosophyNon-prescriptive/conceptual (Issues)– Describes the issues that need to be satisfiedfor a validated method and the minimumstandards Allows for scientific judgment Prescriptive may be too restrictive Cannot conceive & capture all situations– Not intended to be an extensive treatise onmethod validation or analytical platforms4

FDA BMV GuidanceOur feedback (docket, CC V) PrescriptiveConceptual Many comments to include issues we don’t thinkwe need to comment on E.g. Statements about anticoagulants withdiffering counterions5

Bioanalytical Guidances MHLW Draft Guideline onBioanalytical Method Validation (Ligand Binding Assay)Validation in Pharmaceutical Development-2014 FDA Guidance for Industry: Bioanalytical Method validation-2013 MHLW Guideline on Bioanalytical Method Validationin Pharmaceutical Development-2013 ANVISA-Bioanalytical guidance-2012 EMA-guideline on Bioanalytical Method Validation-20111/20126

HarmonizationThere is extensive harmonization of theguidances/guidelines Chromatography, LBA, essentially the same WRIB, AAPS, APA, LOL, EBF, JBF, etc7

Differences .Biomarkers, Diagnostic KitsWhat are the criteria and parameters thatshould be followed?DBS, and other new technologiesshould be crossvalidated with prior standardmethods to interpret dataCross-/Partial validationsWhat are the criteria and parameters thatshould be monitored? In what situations?8

Future Issues .ADCs, LC/MS of large moleculesWhat are the criteria & parameters that are needed forthese methodologies?9

Questions from the JBFQ1. Cross-validation (L122-123)Does “The comparisons should be done both ways.” mean comparing thedata analyzed by both 2 methods using same samples? In case that thereference method is not available, how should we handle this?A little grammatical clarity is needed in the Guidance.Crossvalidation is a comparison of (at least) two methods.QCs and samples should be used, by analysis with both methods.What are the appropriate acceptance criteria?EMA: 15% for QCs, 20% for samplesMHLW: 20%10

Questions from the JBFQ2. Recovery (L238-240)Is Low and High sufficient? We think there’s no need to addMedium.Thank you. We will discuss this during our revision.11

Questions from the JBFQ3. Calibration (L284-295)In case 75% of non-zero standards except the ULOQ met the criteria, is thestandard curve accepted? If it is accepted, the quantitation range would narrowexcept the ULOQ or keep original range ?ULOQ is defined by the highest concentration onthe standard curve. If you dropped the highestconcentration, the next highest concentrationwould define the top end of the calibration curve.12

Questions from the JBFQ3. Calibration (L284-295)What does the sentence “Excluding an individual standard should not change themodel used.” mean?1) When any of the concentration points fell outside the criteria, the calibrationcurve should be re-constructed without that point.2) The calibration curve avoid the need to re-construct. The model (weightedleast-squares regression method, weighting and other) should not change.Correct. Loss of one calibrator should not alter themathematical model used to fit the curve(if it does, that is indicative of a problem with the method).13

Questions from the JBFQ4. QC samples (L399-408)We hope to delete the sentence “e.g., capacity limit of 96-well.analysts”, because this provision makes the pretreatment moretroublesome.This is a difficult issue. Without inclusion of QCson separate plates etc, it is difficult to monitoranalytical performance adequately.14

Questions from the JBFQ5. Carry-over during sample analysis (L438-439)In routine bioanalysis, should the carryover be included into acceptancecriteria of an analytical run as well as accuracy of a calibration curve and QCsamples?Carryover needs to be assessed as part of yourmethod. Ideally, it will be corrected. But if somecarryover persists, you will need to monitor itduring study sample analysis. If it becomesproblematic during a run, you may need to rejectthe run and re-assess/correct the issue.15

Questions from the JBFQ6. Change of counter ion on an anticoagulantBased on the discussion in CC-V, I understand that a change of counter ionon an anticoagulant does not require a partial validation. Please confirm thatadditional stability data is not also required with the ‘new’ counter ion.It does NOT require a partial validation,nor a stability assessment.16

Questions from the JBFQ7. Re-integration (L459-463)Do criteria for reintegration mean rationale for reintegration? Could youpresent example of criteria for reintegration?Yes, the criteria for reintegration is another way ofsaying what is the rationale for the re-integration.A shoulder peak on the analyte could be a legitimatereason, defined a priori, as the basis forre-integrating the peak of interest.17

Questions from the JBFQ8. ISR (L752-783)We hope the number of ISR samples is harmonized among JP MHLW, EUEMA and US FDA.The FDA is agreeable to MHLW and the EMAaccepting 7%18

Questions from the JBFQ8. ISR (L752-783)Does that mean not all PK study should do ISR evaluation? We need toconfirm what studies we don't need to do ISR. Does PD marker need ISR?ISR is intended to be a “spot check” of the method. It should be applied toStudies that support regulatory actions. This would include biomarkers andPD assays, if they were to be used for this purpose.e.g. Studies that need ISRPK study to support labelingBE studiesStudies that may not need ISRearly BE studies changes in formulation (not the final formulation)exploratory studies of PKexploratory studies of biomaker (e.g. Phase 1)19

Questions from the JBFQ8. ISR (L752-783)We think that the acceptance criteria for ISR (20% for small molecules or30% for large molecules) should also be defined based on the difference ofmeasurement principle (Chromatographic assay or Ligand binding assay),not molecular size.Generally, small molecules equates tochromatographic methods (20%)And LBA equate to large molecules (30%)What about LC/MS (chromatography) of largemolecules? –20%?20

Questions from the JBFQ9. InspectionPlease indicate clearly which item will be applied for futureinspection/investigation retrospectively, if there are any. It is unfair toapply some rules against old projects/ studies retrospectively.Some new things aren’t new e.g. LBA criteria,which have been used since 2006.Other things are too new .When the final Guidance is posted, it will describewhen/how things will be in force.21

FDA Guidance Revision-UpdateDraft Guidance posted in Sept 2013Ninety-day comment period-closed Dec 2013Feedback on the Guidance AAPS-FDA: Crystal City V Emails Comments submitted to the docket22

FDA Guidance Revision-UpdateDocket: 640 pages of comments; from around the world. Individuals and consortia Basic types of comments Grammar/clarification/semantics Scientific suggestions/recommendations Unsubstantiated comments23

FDA Guidance Revision-UpdateCompleted database of comments Next: sort and codify Revise text and structure Re-circulate within FDAWhen will this be completed?-Unclear24

ありがとうA RahmanBrian.Booth@fda.hhs.gov25

Bioanalytical Guidances 6 MHLW Draft Guideline on Bioanalytical Method Validation (Ligand Binding Assay) Validation in Pharmaceutical Development-2014 FDA Guidance for Industry: Bioanalytical Method validation-2013 MHLW Guideline on Bioanalytical Method Validation in Pharmaceutical Development-2013 ANVISA-Bioanalytical guidance-2012

Related Documents:

“Bioanalytical Method Validation, A revisit 17: 1551-1557 (Jan,) AAPS Workshop on“Bioanalytical Methods Validation for Macromolecules” (Mar) al., “Workshop on Bioanalytical Methods Validation for Macromolecules: Summary Report” Pharm Res. 2001; 18: 1373-1383 FDA, Guidance for Industry (May) - Bioanalytical method validation - Post

January 1999: Draft Guidance on Bioanalytical Methods Validation was issued by FDA January 12-14 2000: 2nd AAPS/FDA Workshop on ioanalytical Methods Validation A Revisit With a Decade of Progress. Summarised in the 2000 Workshop Report: Shah, V.P. et al., Pharm Res, 2000; 17:, 1551-1557 2001 Guidance on Bioanalytical Methods Validation

Background Current guidelines for bioanalysis in China -Chinese Pharmacopoeia 2000-2010 -Included in BA/BE guideline (1 page) -CFDA guidelines 2005 Objectives of the new guidance in ChP2015 -A separate guidance for bioanalysis -Harmonized with international guidelines (EMA, FDA) -Detailed information -Meet future demands in China

3. ICH Harmonised Guideline, Bioanalytical Method Validation, February 2019, Draft. 4. US FDA Guidance for Industry, Evaluation of Internal Standard Responses During Chromatographic Bioanalysis: Questions and Answers, September 2019. Cultivating reliable data Incorporating IS in bioanalytical methodology is

National Health Surveillance Agency of Brazil (ANVISA) and the FDA Guidance for Industry Bioanalytical Method Validation (2013). The average of recovery was 98.71% in a linear range from 0.1 to 20 μg/mL. The intraday and interday precision and accuracy were within specified limits to bioanalytical methods.

1 Guidance for Industry, Bioanalytical Method Validation (2001) and Workshop/Conference Report – Quantitative bioanalytical methods validation and implementation: best practices for chromatographic and ligand binding assays (Viswanathan et al., AAPS J. 2007; 9 (1) Article 4).

Bioanalytical Method Validation 05/24/18 Bioanalytical Method Validation Guidance for Industry . U.S. Department of Health and Human Services Food and Drug Administration

Abrasive jet Machining consists of 1. Gas propulsion system 2. Abrasive feeder 3. Machining Chamber 4. AJM Nozzle 5. Abrasives Gas Propulsion System Supplies clean and dry air. Air, Nitrogen and carbon dioxide to propel the abrasive particles. Gas may be supplied either from a compressor or a cylinder. In case of a compressor, air filter cum drier should be used to avoid water or oil .